The component XXX appears to be in the middle of a parking sequence" #7357

Discussion options

You must be logged in to vote

This error means that an axis that was parking or unparking itself but the reflectometry IOC was restarted before it was finished.
To resolve this you need to find out where the axis currently is and move it to either completely parked or completely in the beam.

This needs to be done carefully if there are clash conditions for the movement. The positions the component was going to move through can be found in the configuration file, several axes may be involved so make sure you move all axes for the component. The axes should be moved using the table of motors.

Once the component is in the desired place please restart the IOC and it should come back up without a problem.