I would not use MC always to get xray spectra since they were measured already a long time ago. For Xray tube "simulation" one can refer to the CD-ROM published with this 1997 report, part of a publication of the British NHS: Catalogue of Diagnostic X-ray Spectra and Other Data, IPEM Report 78, K. Cranley, B. J. Gilmore, G. W. A. Fogarty and L. Desponds (Electronic version prepared by D. Sutton). The Institution of Physics and Engineering in Medicine and Biology. 1997. ISBN 090418188X This program contains a large set of measured and simulated Xray tube spectra in the medical diagnostic range. See Simon Evan's 1998 review for additional information. Yes, this is a very old program. But still useful. It can be executed in windows 3.11 (!!) See |
Beta Was this translation helpful? Give feedback.
-
This question was originally posted here, and I have copied it over to preserve the discussion: https://www.reddit.com/r/EGSnrc/comments/sh1kwb/x_ray_tube_simulation_in_egs_through_toutor7pp/
The user also asks if it is OK to use
ECUT=AE=0.521